Test/review of DMM Siglent SDM3045X

$
0
0
Forums: 

DMM Siglent SDM3045X

DSC_3373

This is a low end bench meter, but it has many of the function of a high bench end meter, but not the resolution.

DSC_3372

It arrives in a big very solid cardboard box with the DMM well protected inside.

DSC_3375

It includes the meter, a quick start guide, a calibration certificate stating that the meter is in calibration, mains cable, a pair of probes with alligator clips and a usb cable. It do not include any manuals, they must be downloaded.

DSC_3376
DSC_3377
DSC_3378

Standard probe, with good safety.

DSC_3382

The plug is fully shrouded.

DSC_3380
DSC_3381
DSC_3379

Isolated alligator clips to mount on the probes.

DSC_3403

The front has the display, the buttons and the input.

DSC_3383
DSC_3391

There are vent holes on the sides, but the meter do not have a fan.

DSC_3392DSC_3404

The meter can be tilted in two different positions.

DSC_3386

With the handle at the front the meter is ready to be carried.

DSC_3387
DSC_3405

The back, there was a warning to check the mains voltage.

DSC_3388

The input fuse is on the back of the meter.



Display

DSC_3419

This meter is a computer that needs some time to boot and do self test, during that time it shows the brand name. Total time is about 16 seconds.

DSC_3413

Typical display when measuring without any of the more advanced stuff activated. Here I am measuring a voltage.

DSC_3415

Secondary display enabled and showing frequency. Many function on the secondary display will mean a relay has to change mode of the meter. Any combination of VDC/VAC/IDC/IAC can be shown on the main/secondary display, the ACV can also be combined with frequency or period.

DSC_3409

A bargraph can be enabled, it is possible to configure endpoints.

DSC_3411

The statistic can show max/min/average and a bit more.

DSC_3407

It is possible to show data as a chart.

DSC_3408

Or as a histogram with up to 400 bins.

DSC_3418

The collected data can be saved to usb, but it requires a number of button presses:
Utility, Store/Recall, Store settings, Browse, arrow down, Select, Type, Store Data, Yes, Done.

DSC_3420

There is also a help system build into the meter.



Functions

DSC_3401


  • DCV/DCI: DC voltage and current.
  • ACV/ACI: AC voltage and current
  • ohm 2W/4W: Resistance in either two or four wire mode.
  • c/Freq: Capacitance and frequency
  • Cont/diode: Continuity and diode.
  • Temp/Scanner: Temperature with either a thermocoupler or PT100 sensor.
  • Dual/Utility: Select dual display measurement and save/configure/test/update.
  • Acquire/Help: Set trigger and look in build-in manual.
  • Math/Display: Statistic, Limits, DB, relative, bargraph, chart, histogram selections.
  • Run/Stop: Select to measure or not.
  • Single/Hold: Trig a single measurement and enable auto-hold mode.
  • Shift/Local: Access all the blue functions and cancel a remote connection (Local only need a single press on the key).
  • a1a2a3a4: Selection and adjustment in various menus.
  • OK: Confirmation of selection in menus.
  • plus/Range/minus: Change between auto range and manual range and select range.



DSC_3395

The 6 keys below the display are soft keys and the actual function will be listed on the display.
The power button is a soft switch, this means part of the meter is always on and using power.
A short press will turn the meter on a longer press will turn it off.
The led below the button is steady green while on and slowly flashing when meter is off.



Input

The inputs are only CAT II 300V rated, i.e. the meter is not designed for industrial and service on electric installations.

DSC_3402


  • HI: The voltage, frequency, capacitance, diode and ohm positive input.
  • LO: The voltage, frequency, capacitance, diode and ohm negative input.
  • HIsense: Four terminal ohm.
  • LOsense: Four terminal ohm.
  • 10A: Current input


DSC_3389

The trigger in input, this is on of the conditions that can be used to start a measurement. The VM comp will signal when a measurement is done.

DSC_3390

The usb and lan connections on the back can be linked to a computer for remote control and logging.

DSC_3396

There is also a usb connector on the front, it can be used with memory sticks for saving data and configuration.



Measurements

  • Volt and frequency

    • At 100mVrms frequency input range is from 2Hz to 500kHz
    • Frequency counter has a capacitor at the input and ignores any DC offset.
    • 1 VAC is 5% down at 320kHz, rms will not work at this frequency
    • Period is 1/frequency, it do not show high/low period time.
    • Min/max captures in about 200ms on slow speed and 15ms in fast speed, but peak must match sample window.
    • Min/max is based on normal measurements, there is no special mode/circuit to make faster captures, this is standard for bench meters.
    • Meter can show dB with adjustable reference either measured or entered.
    • dBm reference impedance has these settings: 50, 75, 93, 110, 124, 125, 135, 150, 250, 300, 500, 600, 800, 900, 1000, 1200, 8000,
    • Input impedance is 10Mohm on DC, but only 1Mohm on AC with capacitor coupling.
    • Frequency input is similar to AC input.
    • mVDC range can be set to high impedance.
    • The VDC range will toggle a relay, even when not changing range. This may reduce reading speed significantly!

  • Current

    • Current range is protected with a 10A/250V 5×20mm glass fuse accessible from outside.
    • There is also a internal 12A/250V 5×20mm ceramic fuse for protection.
    • uA range has very low burden voltage, it do not use the common 100ohm shunt.
    • Auto range will disconnect shortly when changing from 6mA to 60mA range and 600mA to 6A range.
    • Even in manual range the meter will auto range when a current range is overloaded.

  • Ohm, Continuity, diode and capacitance

    • Ohm needs about 2s to measure 100ohm in default slow and about 0.3s on fast.
    • Ohm is up to 5.0V open and 1.0mA shorted
    • Continuity is quick (About 30ms).
    • Continuity beeps when resistance is below 50ohm, this is adjustable.
    • Continuity is 5.0V open and 1.0mA shorted
    • Diode range uses 5.0V, max. display is 4.000V at 1.0mA, max. current is 1.0mA shorted (Max. display is adjustable from 0 to 4V).
    • Diode range maximum is adjustable, it follows selected threshold, i.e. with 2V as threshold meter will maximum show 2V (Silly).
    • 10uF takes about 2.5 seconds to measure.
    • 9700uF takes about 14 seconds to measure.

  • Miscellaneous

    • Power consumption when on is 9 watt
    • Power consumption when off is 3 watt
    • The meter reach final value in a few updates, extra updates is required when doing auto range.
    • Viewing angle is good.
    • Display update speed depends on settings (5, 50, 150 / second)
    • Bargraph updates will follow the display update, this is normal for bench meters.
    • At 150 readings/second the resolution will be reduced to 3.5 digit.
    • The last 10000 measurement are stored in a internal buffer, this can be presented in various ways, saved to a usb key or downloaded.
    • Weight is 3820g including bumpers.
    • Size is 300 x 260 × 110mm including bumpers and handle.

  • Probes

    • Probe resistance 26mOhm for one.
    • Probe wire is very soft and 98cm long.


1uF

A look at the capacitance measurement waveform.

DMMInputVoltageSweepmVDC

The mVDC range is fairly similar to handheld meters when in high impedance mode.

csv_list

The data format saved to usb in very simple, no time, mode or anything, just the values.

DMMschema

High AC voltage can make DCV show to low value. AC input is capacitor coupled and will not be confused by DC.
The meter has trouble measuring close to 10000uF, it frequently shows overload.
When measuring 10Mohm resistance I can change the result 1Mohm by touching the cables (Using slow speed).



Software

This software is a free download from Siglent.

According to the installation manual you have to download and install NI-VISA yourself before installing this software, I already have this software on my test computer in the newest version and did not do anything.
The software is obvious for all Siglent bench multimeters and have lot of function for a meter with a multiplexer/scanner card, these are not relevant and do not work for this meter and I will not us them..

1s

Starting the software with the meter connected gives a complex looking user interface without much contents.

2s

Plugging the meter in and using “Add instrument” makes it possible to select the meter.

10s

Next time I start the software the meter will be listed and I only have to press connect. Once I had some problems that I could not start the meter after a connect, removing the meter and adding it again fixed that.

3s

Meter is connected and ready, press the start triangle and it is gathering data.

4s

Many of the meter functions are directly available here, but the check is not as stringent as on the meter, I could select dB for DC.
At the bottom is the statistic function.

5s

The bottom can also show the bargraph.

6s

Or a chart

7s

And the histogram.

18s

The temperature function with all selections from the meter available.

17s

It is possible to view the thermocoupler definition. The heading of the table is wrong, it is voltage (millivolt), not resistance.

14s

The gear button beside the meter display has some configuration, including local logging of data.

8s

When the software is running it will collect a local log, these can be viewed and saved.

testexport

The saved data is a very simple format (The actual time format can be configured) and has the usual bug with point and comma. I will be very hard to read this data because comma is used both as a decimal point and as a delimiter.
The filename will match the meters mode, i.e. when measuring DC voltage the filename will include “DC Voltage(V)”.

9s

The viewing of the local logs also includes a chart

15s

Or a histogram. There is a lot of settings on this chart and the window can be resized.

16s

The chart can also be saved to disk or to clipboard.

11s

The “Control” button in the interface opens other ways to use the meter, here a direct SCPI interface.

12s

Fetching of screen dumps. This function is not synchronized with display updates.

13s

A virtual front panel.

19s20s

This interface can scale fairly well and the “DMM Explorer” window can be moved outside the main application.


Tear down

DSC_3423

Four screws at the back bumpers and one at the bottom then the lid was loose, but I also had to remove the handle (Turn it to the right angle and pull out), before I could pull the lid off.

DSC_3430

The front bumpers each has four screws, but it do not open anything up. It is probably for brackets that can be used for rack mounting.

Temp6196

M1: 36.4°C, HS1: 37.2°C
When turned off it is only the transformer that uses power.

Temp6197

M1: 55.3°C, M2: 46.5°C, HS1: 55.7°C
When on the processor and the +/-15V voltage regulators get a bit warm.

DSC_3424DSC_3425

DSC_3426

The meter is in two parts, one for the transformer and one for the electronic. It is without a fan.

DSC_3427

At look down into the electronic section, it has two circuit board. One for the processor and one for the analog stuff.

DSC_3429

The mains input wires goes to a small circuit board together will all other mains wiring.

DSC_3447

The transformer has 3 output voltages, two for the analog board and one for the processor board.

DSC_3431DSC_3432

Back and front of the circuit board section. The connections from the front is soldered to the analog board.

DSC_3433
DSC_3436

DSC_3437DSC_3438

The two BNC on the back share a small circuit board that connects them to a flat cable.

DSC_3440

On this board is the processor (U1: AM3352 Arm Cortex A8), some flash memory (U4: MT29F2G08 2GB Flash) and a DDR RAM (U3: Marked SEC743/K481616/46IBCNA/YBE0270ZC) with a bus termination voltage controller (U15: LP2998). The last of the larger chips (U10: Marked 79CC91G3 / 83848KS0) is probably a network interface.
The power supply is a advanced power management chip (U2: TPS65910), a single switcher (U11) and some linear regulators (U12 & U14: L1117).

DSC_3441

DSC_3442

DSC_3443

DSC_3439

DSC_3444

The other side is mostly resistors and capacitors. There is also a large transistor (Q3) to handle the soft power on for the processor board.

DSC_3448

The analog board do not have any components on the bottom.

DSC_3460

The controller on the analog section is a Lattice PLD (U1: LCMX0640C-3TN100C 640 LUTs & 74 I/Os), it is connected to the processor board with a isolator (U201: ISO7641FC 3000V RMS isolation) and two optocouplers (U706 & U708: Cosmo K1010 5000 V RMS isolation). The analog converter (U511: AD7190 24 bit) is way better than this meter needs, the reference (U407: REF5025K 3ppm/°C) is also very high specification. There is support from some OpAmps and analog switches and +/-15V analog supply (U2 & U3: LM79L15 & LM78L15))
The input resistors are build from many resistors in series. The Sense input has 4 resistors (R301, R313, R314, R315: 4×29.9kOhm & R300, R310, R311, R312: 4×29.9kOhm) with diodes at the end (D300 & D301) for each terminal. There is also a 100kOhm input chain (R302, R303, R304, R305, R306, R307, R308, R309: 8×13K), the reason for that many resistors is probably power handling during overload and a 10Mohm chain (R73, R74, R75, R76>: 4×2.5Mohm).

DSC_3456

The input area has 5 mechanical relays for range selection. Below the white relay is the current shunts for 10A (R408: 0.01ohm) and uA/mA (R407: 1ohm) and a OpAmp (U300: TL072C). There is also the 1000 volt input capacitor, two MOVs and two spark gaps.
In addition to the fuse on the backside there is another one here, it is ceramic, but not a high voltage HRC fuse.

DSC_3451

Below the metal shield is OpAmps, analog switches and the RMS converter (U711: ES636 0.1%)

DSC_3454


DSC_3461DSC_3462

Zoom in on part of the analog board, all IC numbers can be read here.



Conclusion

This is not a top end bench meter with “only” 4½ digit, but the meter works the same way as they do, with user interface, communication, relays/analog switches to handle all range changes. This meter is not a handheld meter with a mains power supply build in, but a real bench meter.
Generally the meter works exactly has expected, that means fairly good. I like that the current input is auto ranging from uA to A. The support for thermocouples is not present in all bench meters, but here it is, together with a minimum PT100 support (It is missing 3 & 4 wire support).
The software is good for a multimeter software and that it is free makes it even better.

I believe you get a lot of bench meter for the money with this meter, especially if you want to use the computer connection.



Notes

Siglent has 3 meters in this series:
SDM3045X: 4½ digit
SDM3055X: 5½ digit
SDM3065X: 6½ digit, can be ordered with a build in multiplexer.

The layout and user interface is very similar to Keysight meters

How do I review a DMM
More DMM reviews