Test/review of DMM Owon B35T

$
0
0
Forums: 

DMM Owon B35T

DSC_8573

Own has a couple of multimeters in this “35” series with some differences. The “B” models has Bluetooth, the “T” models has true RMS and the “+” model has memory for offline logging. I look at the model with Bluetooth and true RMS, but without offline logging.

DSC_8488DSC_8490
DSC_8489
DSC_8491
DSC_8492

The meter arrived in a brown cardboard box with a minimum of text on it.

DSC_8562

Inside was a pouch with everything in it.

DSC_8564

The box/pouch included the meter, probes, alligator clips, a thermocoupler, transistor tester adapter, pouch, quick guide, specifications, QR code for link to app.

DSC_8566
DSC_8567
DSC_8568

Probes are marked 1000V 10A and the tip cover is marked CAT III, without the cover there is a CAT II marking.
The tip cover is screw on/off, the alligator clips are also screw on/off.

DSC_8570

The plugs has are fully shroud, but a bit on the short side.

DSC_8569

The transistor tester, thermocoupler, capacitor adapter. This way the transistor tester is safe.

DSC_8571
DSC_8572

A cheap thermocoupler with banana plugs, but notice the black plug is on the red wire (There is a reason for that).

DSC_8577DSC_8584

DSC_8574

DSC_8575DSC_8576

DSC_8579

I do not like the plastic for the battery lid, I have the feeling that it is a hard plastic that easily breaks.




Display

DSC_8581

The above picture shows all the segments on the display.
The meter do not use all segments.

DSC_8582

Normal DC voltage with voltage and bargraph.



Functions

DSC_8585

Buttons:


  • Select: Select between DC and AC and modes in the ohm position.
  • Range: Switch to manual range and select range, hold down to activate automatic ranging again.
  • Hz/Duty: Select frequency and duty cycle in AC modes and frequency mode.
  • Max/min: Capture maximum and minimum values, press to display one or the other, hold down to exit mode.
  • light: Press to freeze/unfreeze current reading, hold down to turn on background light,
  • rel: Press to use relative mode (REL), hold down to enable/disable Bluetooth.

REL and Max/min disables auto range.

Rotary switch:

  • Off: Meter is turned off.
  • V: Measure DC and AC voltage, DC includes a mV range, in AC it is possible to select frequency.
  • mV: Millivolt range, has two mV ranges for both DC and AC, in AC it is possible to select frequency.
  • ohm: Resistance, diode, continuity and capacitance
  • Hz%: Frequency, this support up to 10MHz.
  • hFE: Can check transistors with the supplied adapter.
  • Temp: Temperature.
  • uA: The uA range, in AC it is possible to select frequency.
  • mA: The mA range, in AC it is possible to select frequency.
  • A: The A range, in AC it is possible to select frequency.




Input

DSC_8586


  • A: High current, maximum current is 20A
  • mAuA: The lower current ranges and input for thermocoupler (Note polarity is reversed) and transistor tester.
  • CON: The common terminal for all ranges.
  • xxx: All other ranges.




Measurements

  • Volt and frequency

    • Frequency input do not require a zero crossing
    • At 100mVrms frequency input range is from 0.3Hz to 1.6MHz
    • At 1Vrms frequency input range is from 0.3Hz to 14MHz
    • Duty cycle works from 5% to 90% at 10kHz with 1Vpp, precision is within 5.7
    • Duty cycle works from 1% to 99% at 10kHz with 2Vpp, precision is within 1.2
    • 1 VAC is 5% up at 70kHz (RMS will not work at the frequency).
    • Max/min needs about 200ms to capture a voltage, it may needs multiple pulses.
    • Input impedance is 10-11Mohm on DC and AC, except mVDC that is high impedance
    • mV in mV position has 10Mohm input impedance up to about 0.8V, then it drops to 100kOhm at 5 3olt and 2kOhm at 20 volt.
    • Meter has two mVDC ranges, one in the V position and one in the mV position
    • There is an audible warning when voltage is too high on mV (600mV) or V (750/1000V) range.
    • Frequency input is 10Mohm up to 12 volt then it drops to 2kohm at 20 volt
    • Rated overload protection on V is 1000VDC / 750VAC
    • Rated overload protection on mV is 250VDC or equivalent AC

  • Current

    • Overload protection in uA and mA: 1A/250V 5×20mm glass fuse
    • Overload protection in A: 20A/250V 5×20mm ceramic fuse
    • Measuring above 10A must only be performed for a short time.

  • Ohm, continuity, diode and capacitance

    • Ohm needs about 2.9s to measure 100ohm
    • Ohm is 0.64V open and 0.34mA shorted
    • Continuity is quick (80ms).
    • Continuity beeps when resistance is below 21ohm
    • Continuity is 0.64V open and 0.34mA shorted
    • Diode range uses 2.6V, max. display is 1.999V at 0.3mA, max. current is 1.2mA shorted
    • 10uF takes about 2.7 seconds to measure.
    • 4000uF takes about 30 seconds to measure.
    • Rated overload protection is 250VDC or equivalent AC

  • Miscellaneous

    • Polarity of temperature input is reversed, the plugs on the thermocoupler is also mounted reverse.
    • Current consumption of meter is 2.3mA and 3.2mA depending on range (AC adds 0.5mA). AC with backlight is 20mA, AC with Bluetooth is 13mA (Everything together is about 28mA).
    • Meter works down to 1.8V where it turns off, battery symbol show at 2.3V.
    • The meter reading is stable at any battery voltage.
    • Backlight fades with voltage and is very dim at 2.5V
    • Bluetooth works until meter turns off due to low battery.
    • Bluetooth cannot automatic recover a broken connection.
    • The meter needs a couple of updates before the reading is fully correct.
    • Viewing angle is good
    • Display updates around 3 times/sec.
    • Bargraph updates 30 times/sec
    • Backlight will turn off in 47 seconds
    • Will automatic turn power off in about 15 minutes
    • Standard probes cannot be fully inserted into sockets on meter, but make connection, banana plugs can be fully inserted.
    • Weight is 335g without accessories, but with batteries.
    • Size is 185 × 88 × 45mm.

  • Probes

    • Probe resistance 31mOhm for one.
    • Probe wire is soft and 89cm long.


10uF

A look at the capacitance measurement waveform.

DMMInputVoltageSweepHz

Frequency input resistance.

DMMInputVoltageSweepmVDC

Input impedance on mV range, on V range the mV input has high input impedance, but not here.

DMMschema

The 40nF range has a -0.2nF offset, using REL the offset will be +0.2nF instead (The negative offset will only be visible as too low readings when measuring a capacitor).
Battery time is without Bluetooth, with Bluetooth it is around 200 hours.
Large DC voltage will block for AC readings.



Software Android

The Bluetooth phone software must be downloaded from Owon site and requires access to phone information, contacts, location and storage. I disabled the phone and contact access and the application still works.

Screenshot8s

The software I got was V1.2.9 and I downloaded the Bluetooth 4 version. The program is called “MultimeterBLE” on Android.

Screenshot1s

The typical meter display, the blue fields will activate buttons on the meter (it is still possible to use the button on the meter).
The microphone will enable English voice annotations of the value, the language is not perfect but understandable. The menu besides the microphone is used for meters with internal logging.
The software will only work in landscape mode.

Screenshot7s

Here max/min has been activates, it is shown, but could have been more obvious.

Screenshot2s

The software can support more than one meter, I only had one meter and have not tried this function.

Screenshot3s

The logging mode, it shows a curve and a table, this data is also collected when showing the other display.
The chart has some problems: It do not change scale between mV and V, i.e. 600mV is shown as 600 and 2V is shown as 2 on the chart, this can occur in the same chart! This means it may be necessary to use the RANGE switch to get consistent charts.
I did not find a way to stretch the chart or table to full screen width. The chart and table will not reset if the range switch is turned, it will just continue to add data.

Screenshot4s

It is possible to zoom in/out on the curve, either with the buttons in the gray area or with two fingers.

Screenshot6s

On this screen the menu has function to save the data and share the saved data. I shared it with an email program and received the email on my PC.

Screenshot5s

The settings can be used to adjust the sample interval and to set a time limit on how long to record the data. The “Fill” option will add a blue shading below the curve.

data

The data is saved in a csv file using US format and without any headers, this means I cannot directly import it into Excel (I need to specify format each time).

The protocol is documented on Github and there is also a Linux program for communicating with the meter.



Tear down

DSC_9606

I had to remove four screws to open the meter

DSC_9607

DSC_9608

The fuses have a cover on them, the 20A one has a bit heat damage from my 20A test.

DSC_9609

DSC_9610

Five more screws must be removed to get the circuit board out.

DSC_9611

DSC_9612

And four screws to remove the LCD display with backlight.

DSC_9616

The two fuses are rather small, but there is space for larger fuses on the circuit board, this probably means a meter bought from a EU distributor will have a large fuse. The 20A shunt (R38) is rather small, this is probably the reason for the time limit. The mA shunt is two resistor (R39 & R40: 2×1ohm) in parallel, one on each side of the circuit board. The uA shunt is the next resistor (R43: 49.9ohm). These resistors have about half the usual resistance, this is very nice for burden voltage. The uAmA shunts are protected by some diodes (D11, D12, D13, D14),
The voltage input has two paths, both with a PTC (R50 & RT1: 1kOhm), the main voltage input goes to two resistors, bypassed with two capacitors and a resistor (R22 & R33: 5Mohm, C45 & C49 & R80: 100kOhm). This improves the frequency response. The ohm range uses two transistor pair for protection (Q3, Q4, Q5, Q8), together with the two PTC’s. Frequency uses one transistor pair (Q4 & Q5).
I expect many of the diodes is used to encode the range switch for the multimeter IC.
The main meter IC is from Semic (U1: CS7729CN-001) with an external RMS converter (U10: AD8439JCPZ). The Bluetooth interface is a module (Chip is CC2541).
The meter has a couple of trimpots (V2, VR3, VR4, VR5, VR6, VR7) and a trim capacitor (C39).

DSC_9617

DSC_9618DSC_9620

DSC_9619

DSC_9613

On this side is all the pads for the range switch, the buttons and the LCD connection. There is also one of the mA current shunt resistors (R40) and protection diodes (D12 & D14). The main voltage input resistor is two resistors (R22 & R33).

DSC_9614

DSC_9615



Conclusion

As is often the case the CAT rating is way optimistic, 250V 20mm glass fuse do not belong in a 1000V CAT III meter. The lack of MOV’s and the small PTC’s without series resistor will also limit the ability to handle high voltage transients.
This meter has all the normal ranges for a competent meter and has a nice fast bargraph, but lacks any advanced functions. The input protection is fine for occasionally mains use. The lower shunt resistance (Burden voltage) in current ranges is nice, Bluetooth can be used for logging data at fairly short ranges



Notes

How do I review a DMM
More DMM reviews

My website with reviews of many chargers and batteries (More than 1000): https://lygte-info.dk/