Multimeter guide
This guide is a simple (but long) guide for many details on multimeters, it do not tell how to make measurements, but explains about the different function on the multimeter. This guide has lot of pictures, each showing a small part of a multimeter to illustrate the functions and differences between meters. Most of the article is about measurement function, but at the end I will also look at a few physical aspects on the meter.
This guide can be very useful to make a list of functions you want in your next multimeter.
Contents
Minimum, maximum and average of a value
Peak or fast minimum, maximum
Freezing the display
Manual selecting range
Selecting function
Background light and flashlight
Frequency and duty cycle
Relative measurements
Auto power off, can it be disabled?
Dual or triple display
Bargraph, sometimes a fast estimate of value
Voltage measurement
Current measurement
Ohm, continuity, diode and capacity measurement
Temperature measurement
Non contact voltage or electric field detection
Low-Z, getting rid of ghost voltages
Low pass filter for variable frequency drives
Turning the meter on or off
Input terminals
Tilting bale or kickstand
Batteries and fuses
Probe holders, need an extra hand
Hanger
Minimum, maximum and average of a value
Many multimeters have a function to record minimum and maximum value, it is fairly slow (0.1 to 1 second), this is necessary to work for AC and not record random noise pulses. This function will usual work on most ranges on the multimeter, except capacity.
Depending on meter is may be possible to show actual value while recording minimum and maximum. Some of the better meters will also calculate a average while recording minimum and maximum. Meters with dual or triple display may use the extra display(s) for some of this.
To activate the function press the button, more presses on the button will change between minimum/maximum/average/actual depending on the meter. To end recording hold down the button. This function will nearly always disable auto ranging, i.e. remember to manually select the correct range before starting min/max collection.
Most brands agree on calling this function something with min/max or max/min, but Brymen desided to call it REC As long as MAXMINAVG is displayed, the display will show the actual value. When only one of the indicators is displayed, it is that value.
Other meters do also use that notation.
This meter is using the secondary display to show minimum and maximum values.
Peak or fast minimum, maximum
This function can record fast transient, like current while a capacitor charges up (If it is supported in current ranges). It often need around 1ms (0.001s) pulse to record, but it varies with meter. This is fast enough to record maximum and minimum in a mains power sinewave. The function may only be supported in AC modes, but that is no problem because it usual records both maximum and minimum (I have seen exceptions to this). This function will nearly always disable auto ranging, i.e. remember to manually select the correct range.
This function can not replace a min/max, the fast capture means it is much more susceptible to noise and transients.
A PEAK label is added to the display when recording peak values.
This meter do not have a special peak indicator but uses the Hold indicator together with min/max.
Here a P is added before the min/max labels and the value is shown on the secondary display.
Brymen uses CREST instead of peak and add a C to min/max.
Freezing the display
This function freezes the the display, this can be used to move the meter to a better place for reading the value or writing the value down. Just about any meter has this not very useful function.
One press will freeze the display, next press will resume updating again.
Some meters has a auto-hold or touch-hold or something like that, it will freeze the display when measuring a stable value. This makes it possible to concentrate on the probe and when the meter beeps, remove the probes and read the meter. No button presses is required when doing this.
Usual one of these 3 indicators are used to show that the display is frozen.
Manual selecting range
On many auto-ranging multimeters it is also possible to manually select range, sometimes it is even required for a specific ranges (Like mV AC).
MIN/MAX, PEAK and REL will usual select manual ranging, but not on all multimeters. When canceling the function the meter will usual return to automatic ranging again.
Press the button once to lock the current range, press again to step through the different ranges, actual range can usual be seen on the point. Hold the button down to get back to auto-ranging.
This is sometimes used with varing values to avoid the meter changes range all the time. If a lot of resistors or capacitors with about same value has to be checked, manually selecting range will make the meter faster.
Usual there is a AUTO text somewhere on the display when using automatic ranging.
The MANUAL text is not always used.
Selecting function
On most multimeters there are more than one function for each rotary switch position, this button is used to select between them or at least some of them.
The color coding of the button often match the some color coding for the markings around the rotary switch.
Hz and duty cycle is sometimes select with this button sometimes with a Hz button.
In this document I will call the button SELECT.
When the blue switch is pressed the blue function is selected.
Here it is yellow and there are arrows to help with the sequence.
Not everybody believes in arrows. This meters starts in AC mode, the = is orange, i.e. SEL must be pressed to select it.
Here it is not possible to see what initial selection is.
Here it is not even possible to see what the SELECT button do (Southwire had to be different).
Each press will move to the next function for that rotary switch position, usual moving the rotary switch or turning the meter off will reset to the initial function again (Brymen and a few other remembers last selection).
Background light and flashlight
Many multimeters have a light that can illuminate the display, sadly it is often implemented in a way that makes it very awkward to use. On many of the cheaper multimeter it is only on for 15 seconds, then it turns off automatically and it requires a few seconds to turn it on. A few meters uses a light sensor to turn the background light on when it is dark.
The first light button requires a single press to turn the light on, the two next must be held down.
The background light on this multimeter works different: usual the background will turn on due to a light sensor near the display, i.e. when it is dark around the display the background light turns on. The button disables this function and the background light will stay off, until the meter has been turned off.
Some lights has a build in flashlight, that may be able to illuminate the place you are measuring (May require an extra person to hold the meter). This flashlight either has its own button or it turns on together with the backlight.
Flashlight placement is either top or back of the multimeter.
Frequency and duty cycle
The SELECT button may select this, but often an extra button is used for selecing frequency and duty cycle measurements.
It varies where this button can be used, but it is in one or more of these ranges:
- V AC – This is nearly always supported.
- V DC
- mV AC – If the meter is missing a Hz mode, this is usual the range with support for the highest frequencies.
- mV DC
- Hz – In the Hz mode it is often the Hz button that select duty cycle, not the SELECT button.
- uA AC
- uA DC
- mA AC
- mA DC
- A AC
- A DC
The meter may or may not support showing frequency for DC pulses (i.e. pulses that do not cross the zero line) and this may not be releated to support for frequency in DC ranges. Adding a capacitor in series with the multimeter or connecting minus on the multimeter to a voltage a bit above ground will make it possible to handle this and measure duty cycle at drive signal levels (0-3V, 0-5V, 0-12V etc.).
Pressing the button will select Hz, next press duty cycle (Not all meters have duty cycle) and third press back to the main function again.
The Hz selection on the rotary switch will usually be designed for low voltage and often support fairly high frequencies (i.e. MHz). The input impedance will often drop to low kOhm above a few volts.
Relative measurements




This is a very common function, it is used to “null” out a value and show measurement relative to that. Values to null can be the ohm with shorted probes, but it could also be a unloaded voltage, then the meter will only show the voltage drop when load is applied.
Meters with dual display may show the stored value or the full measured value on one of the displays.
This function will often change to manual ranging on the meter.
Auto power off, can it be disabled?




Nearly all meters will automatic turn power off after some time, sometimes they are smart and disables it when measuring a value or when running max/min function.
If the meter shows that auto power off is active, it may have a function to disable it. This is usual done by holding one of the buttons down while the meter is turned on (The button must be held down a few seconds). If the symbol is not shown, then the function is disabled.
Dual or triple display

Some meters can show more than one value at a time. On some meters you have to use the SELECT or a special DUAL button to select reading for the secondary display, but in many cases it is automatic.








Bargraph, sometimes a fast estimate of value
Meters may have a line at the bottom that shows the value, on better meters this line will update very fast. Some meters cheats with the bargraph and grouped segments, i.e. it looks like it has many segments, but they turned on/off in groups reducing the real resolution.




Voltage measurement
Voltage measurement can cover from one to four or more position on the range switch, it includes the normal voltage measurement, but also some special functions.
- V DC – DC voltage up to maximum voltage supported by the meter, that is usual from 600V to 1000V
- V AC – AC voltage up to maximum voltage supported by the meter, that is usual from 600V to 1000V. This may be lower than DC voltage, due to peak voltage being 1.4 times higher than display reading.
- V AC VFD/LowZ – AC voltage with some special function activated, see below.
- mV DC – DC voltage below 1 volt, sometimes this is a separate range, sometimes it is part of the V DC range.
- mV AC – AC voltage below 1 volt, sometimes this is a separate range, sometimes it is part of the V DC range. When part of V AC it may require manual range selection to select it.
In addition to the standard voltage, it might be possible to select some extra function with SELECT or Hz buttons.
- AC+DC – Combined rms voltage for AC and DC
- Frequency – Frequency and duty cycle
- Low pass filter (LPF) – Low pass filter, used on variable frequency drives (VFD’s) to remove high frequencies.
- Temperature – A thermocoupler is very low voltage and the selection may be placed on the mV DC range.









Current measurement
A good multimeter has 6 AC/DC current ranges, two uA, two mA and two A. This usual requires two or 3 positions on the range switch (mA uA requires a switch, A uses another terminal and do not need a extra switch position). Good meters will also have some way to warn the users if they select voltage with probes connected to some current range, to avoid a blown fuse.
Just because the meter has a A on the range switch do not means it has two A ranges, some meters only has the 10A/20A range, not the xA range.








Lets also take a look at the input terminals for current. Usual there is on for A and one for uA and mA, but a few meters move the uA and mA to the voltage input, this can be problematic due to the increased risk of measuring volt on a current range.







Ohm, continuity, diode and capacity measurement
These four function has one thing in common: The meter must supply some current to do the measurement.
This can be up to about 3V and 2mA, but will often be considerable lower, depending on function. Bench meters may use up to 10V and a single multimeter has a 15V diode mode.
- Ohm – Measuring resistance, lowest range may have a resolution from 1ohm down to 0.001ohm (1mOhm), depending on meter, highest range usual goes into the Megaohm, sometimes up to more than 50Mohm. A few meters support up to Gohm, but that is usual measured as ns (nano siemens: 1000000000/ohm).
- Continuity – Audible resistance check, this range uses one of the low ohm ranges (Not always the lowest and test current/voltage may be slightly different) and will sound a buzzer when the resistance is low (Often below 50ohm, but varies with meter and may even be selectable). In addition to the buzzer some meters may also flash a light when continuity is detected.
This is useful to trace wires, because you do not have to check the display, but can use the sound and a good continuity mode is very fast. This means you can pull the probe across a couple of connection and the beep will sound when you hit one with a matching connection. - Diode – Test forward voltage of a diode, some meters uses up to 3 volt and can test white leds, others only use 1 volt and cannot test any leds only regular diodes. If working with white leds a meter that can show 3V in diode mode is very useful.
- Capacity – Measure capacitors, but the low and high end varies significantly depending on meter.
- ohm
The nS is a special ohm mode that measures 1000000000/ohm and can be used for very high resistance. - Continuity
- Diode
It looks like somebody forgot the diode symbol for the second one. - Capacity
It will be nano (n), micro (u) and sometimes milli (m) farad that is shown as unit.









Temperature measurement






Many multimeters can use an external temperature sensor called a thermocoupler. This is a very simple temperature sensor that can cover a wide temperature range, but has a fairly low signal level.
A few meters can handle two thermocouplers at the same time and may be able to show the difference between them, this can be very useful when working with ventilations and airconditions systems.
The thermocoupler is fairly linear for temperature above about -40°C, for this reason many meter either has a lower limit around this or shows large errors at lower temperatures. There is also meters that has compensation and will show correct at lower temperatures.






Non contact voltage or electric field detection





This function can be used to check for mains voltage without using probes, just by holding the meter near to the wires, at least in theory. The sensitivity depends on how you hold the meter and a wet surface or wire will block for this function.
Often both buzzer and a led will be used to notify about a electric field in addition to the display.
The function can be useful for checking what wire is live or in some cases tracking a live wire, just never assume that everything is safe, just because the NCV do not show anything.
The function is usual called NCV and shown as EF on the display, but as can be seen above the selector can also be named EF or even Vsense.




Low-Z, getting rid of ghost voltages





This function is mostly for mains work. Due to the high input impedance (10Mohm) of multimeters two mains wires in a cable or in a pipe will both show voltage when one has voltage and the other is unconnected. This is because there is some capacity between the wires and a small current will leak. Using the Low-Z mode on a meter will put enough load on the wire to show these leaks as safe voltages, but any real voltage (I.e. with enough current to drive loads) will still be measureable.
This function will usual switch the meter to the highest voltage range and sometimes to AC mode. It is often implemented as a PTC, that may be down to about 2kOhm when cold and will increase to maybe 100kOhm very fast when mains voltage is applied. This function may not work with low voltages.
The 400kOhm is slightly different, it simply connect a resistor across the voltage input when held down and do not care about what range is selected.
Most of the time the display will be in V AC mode, but here are a few exceptions:


Low pass filter for variable frequency drives




When measuring on motor drives, also called variable frequency drives (VFD) the voltage will contain many harmonic frequencies. To avoid they skew the measurement some meters has a function to filter higher frequencies out. This is usual only required on high end meters, other meters do not have enough bandwidth in the first place, i.e. there is a low pass filter always active.
The function is usual activated with the SELECT button and will show a small icon in the display:




Turning the meter on or off
All meters needs a power switch, usual it is on the range switch, but a few meters use a button, usual an electronic switch. Neither soft buttons or auto power off will usual not turn the meter fully off, but the current may be negligible.



The most common off location is with the rotary switch all the way to the left.


If the meters has a Low-Z mode it will often be placed below the off position, this makes it fast to select this position when turning the meter on.

Some meters has a off at both ends of the range switch, this can reduce the wear on the range switch if the meter is mostly used for current.


Another way to do that is to place off at the center of the range switch. This will also mean less turning of the switch to select range.


A few meters use a button to turn on off, for better meters it is a electronic switch.
Input terminals


Usual multimeters have the terminals placed in a standard way, like shown above, but there are exceptions.


They do not need to be on a line.

Some multimeters can electronically switch current ranges from uA to A and only need one current terminal. The voltage and current inputs are also swapped on this meter.


Then there are meters with less current ranges.

Here the common and voltage input are swapped and the capacity terminals are mA and V input.

This looks standard, but the distance between COM and voltage input is not 19mm, but smaller. This means modules to plug into multimeters will not fit.

There is also a point about how well standard probes fits into the terminals. The above picture shows a fully inserted probe.

On this small multimeter the sockets are not deep enough to fully insert the standard probe.
The probes supplied with the meter will go fully into the socket.

On this multimeter there is some probe detection that prevents fully inserting the standard probe.
The probes supplied with the meter will go fully into the socket.
Tilting bale or kickstand



Sometimes it is nice to get a bit of angle on the multimeter, instead of having it lying flat.

Most meters have a tilting bale or kickstand on the back, this one with rubber tips, that makes the meter more stable when using it.



Some meters have more positions for the kickstand, making it possible to select the best angle or even use it to hang the meter somewhere.


Another way to make more angles possible.

The kickstand can be part of the multimeter or part of the cover.
Batteries and fuses


All multimeters needs batteries and most also needs fuses. The batteries are usual 9V, AA or AAA batteries, for very small meters coin or button cells may be used. A single 9V or from two to 6 AA/AAA batteries, battery lifetime is from below 100 hours to above 1000 hours depending on the meter and battery.
The fuses need for high current and voltage are HRC (High Rupture Capacity) fuses that can break many thousand ampere. These fuses are ceramic (or better) and is filled with sand. The large DMM44 fuse is of that type.
The glas fuse and small ceramic fuse can probably handle normal mains voltage after a breaker or fuse and are basically for hobby meters.
And then there are the fuses in between, ceramic and hopefully filled with sand.
It is always a good idea to check if the fuses are rated for at least the same voltage as the meter, they must be able to break the circuit if you put the probes in ampere and starts measuring volt. This is the worst case scenario and can be fatal if the fuse cannot break the current in a high current situation.


A typical meter, one screw and there is battery access, but two more screws must be removed before there is access to the fuses.


One more with this two level access.

This one is smarter, here is two different lids, one for batteries and one for fuses.


This meter is a bit difficult, before the battery can be replaced the cover must be removed and again the meter must be opened to replace fuses.


Here the battery is easy to replace, but for fuses the rubber cover must be removed and the meter opened.


Fluke is using a single screw that must only be turned half a turn with a coin and there is access to both batteries and fuses. This meter is a high end meter that uses a lot of power.
Probe holders, need an extra hand
Many multimeters have probe holders build into the body, they can be used when storing the meter in a toolbag, but also when using the meter.

I have never found a good way to pack the wires.


Need to hold two probes and a multimeter, easy with one of the probe holders.

Sometimes you can be lucky and probe wire length match exactly with the probes in the holders.

This meter is too short the sharp end of the probe is poking out.


But the flashlight is very useful in dark places and aimed the same was as a probe in the holder.
Hanger
Many meters has a way to hang them, either from a hook/nail/screw or on some metal.





A lot of meters has this kind of holes on the back, it somtimes be used to hand the meter on a hook/nail/screw, but it can also be used with a magnetic hanger.

This strap is a magnetic hanger, at one end is a powerful magnet with plastic coating.
Sometimes meters will not fit a standard strap, but only a brand specific magnetic hanger.

Here I have mounted it on a multimeter


And the meter can hang on my fridge, metal shelves, electric distribution board in metal, machines. The other meter included the magnetic hanger.
I have also seen a few meters with a magnet build into the back of the meter.
Other articles about multimeters
Multimeter and component testing
Multimeter probes
Multimeter and voltage measurements
Multimeter and pulsed DC current (PWM)
Tolerance specifications for multimeters
Clamp meters magnetization, demagnetization and offset
Multimeters and current measurements
Multimeters and thermocouples
Multimeter design
Multimeter protection and safety
Multimeter reviews
Multimeter selection table
My website with reviews of many chargers and batteries (More than 1000): https://lygte-info.dk/