Test/review of DMM UNI-T UT33A+

$
0
0
Forums: 

DMM UNI-T UT33A+

DSC_4244

This is a fairly simple and small DMM, but has a nice selection of ranges.

DSC_4007DSC_4009DSC_4010DSC_4011

DSC_4012

The box is for all the UT33 meters with a checkmark for the actual model and a table on the back for comparing the models (in Chinese).
If you want all the function, you must buy all the models, none of them can everything.

DSC_4241

In the box was the meter, probe leads and a instruction sheet in both Chinese and English. The instruction sheet is for all four models in the series (A, B, C, D).

DSC_4252
DSC_4253

The standard probes are about normal size, but rating them for 10A is on the optimistic side.

DSC_4254

The plugs are fully shrouded.

DSC_4247DSC_4261

With the meter lying flat it is possible to turn the switch, but on the tilting bale it will slide around.

DSC_4245
DSC_4246
DSC_4248
DSC_4249
DSC_4250
DSC_4251
DSC_4256




Display

DSC_4257

All the segments are shown during power on, not all are used by this meter.

DSC_4258

Typical display during usage, it will show the number and selected measurement



Functions

DSC_4259

Buttons:


  • Sel/rel: Used to select AC in mV and current ranges and to zero capacity range.
  • Hold/Backlight: Will freeze the display reading, until pressed again. Hold down to turn backlight on.

Rotary switch:

  • Off: Meter is turned off
  • VAC: AC Voltage
  • VDC: DC Voltage
  • mV: Milli volt AC and DC, use SEL to select.
  • ohm: Resistance range.
  • Cont: Continuity range.
  • diode: Diode range.
  • cap: Capacitance range, use REL to zero before measuring low values.
  • uA: Micro ampere range, use SEL to select AC.
  • mA: Milli ampere range, use SEL to select AC.
  • A: Ampere range, use SEL to select AC.



Input

DSC_4260


  • 10ADC: 10A DC current input.
  • COM: The common terminal for all ranges.
  • xxx: All other ranges, including mA. It is always problematic when a current range shares connector with a voltage range, if the switch is in the wrong position the fuse will blow (at least). This will only affect current, voltage will still work.



Measurements

  • Volt and frequency

    • 1V AC readings is 5% down at 2.5kHz
    • Input impedance is 10-11Mohm on DC and AC
    • Input protection is 600V AC/DC

  • Current

    • 10A range is fused with a 10A/250V 5×20mm fuse
    • mA range is fused with a 0.2A/250V 5×20mm fuse
    • 10A range has an audible alarm at 10A or above.

  • Ohm, continuity, diode and capacity

    • Ohm needs about 3.5s to measure 100ohm
    • Ohm voltage is 1V open and 0.31mA shorted
    • Continuity is fast (About 20ms).
    • Continuity beeps when resistance is below 31ohm.
    • Continuity is 2.1V open and 0.31mA shorted
    • Diode range uses 2.2V, max. display is 1.999V at 0.07mA, max. current is 0.9mA shorted
    • 2000uF takes about 9 seconds to measure.
    • Overload protection is 600V AC/DC

  • Miscellaneous

    • Current consumption of meter is 1.2 to 1.6mA (15.5mA with backlight).
    • Display gets unstable about 1.6V, but it might show overflow before, battery symbol show at 2.4V.
    • Readings will be correct down to 2.4V, below that they can be high or display can show OL
    • Backlight works down to 1.6V
    • Viewing angle is good, except from the top
    • Display updates around 2.5 times/sec
    • Backlight will turn off in about 30 seconds.
    • Will automatic turn power off in about 15 minutes.
    • Standard probes fits in the meter.
    • The meter usual need a couple of display update to reach the final value.
    • Weight is 208g without accessories, but with sleeve and batteries.
    • Size is 135 × 77 × 47mm

  • Probes

    • Probe resistance 72mOhm for one, that is fairly high.
    • Probe wire is a bit thin, the are 89cm long.



DMMschema

The meter is a true-rms meter.




Tear down

DSC_5512

I had to remove two screws, the battery cover could stay on.

DSC_5513

DSC_5514

The circuit board is shaped to fit the box and uses ceramic fuses.

DSC_5515

I had to remove 3 screws around the range switch to get the circuit board out.

DSC_5516

DSC_5517

DSC_5518

And four screws to remove the lcd display, the backlight is soldered to the circuit board.

DSC_5519

There is no parts on this side, only pads for switches and lcd display and a marking of where ranges.

DSC_5520

The resistor for the uA range (R7: 99ohm) and uA range (R8: 0.99ohm) is protected by two diodes (D1 & D2: M7). The main input shunt is two resistors (R1 & R2: 2×5Mohm), there is also a secondary input chain (R4 & R5: 2X499kOhm). Ohm output is protected by a PTC (PTC1) and two transistors (Q1 & Q2).
The multimeter IC is a COP (Chip on board) that is bonded directly to the circuit board and then covered with some black stuff.

DSC_5521

DSC_5522DSC_5524

DSC_5523

DSC_5525



Conclusion

The CAT rating do not match the fuses used, i.e. it is not correct, but at least it use fuses and fairly good ones.
The meter has a nice selection of ranges, but is missing frequency and min/max functions for a general purpose meter. This is the best meter in the UT33+ range, but the other meters in the range each has one function this meter is missing (Battery test, temperature, NCV).
I was a bit surprised that this meter is a true rms meter.



Notes

UNI-T do often make rebranded meters, i.e. it may exist with other names on it.

How do I review a DMM
More DMM reviews